I'm looking for any information on stand alone and portable medical air compressors or medical gas delivery systems that don't require a centralized medical air compressor. Currently, these are the only medical air compressors I know of. There has to be a smaller alternative for compressed medical gas outside of scroll air compressors and giant reciprocating compressors.
With regards,
Is this for patient care? as medical air in NFPA 99 is limited to patient care?
Something like the attached can not be used for NFPA 99 Category 1 applications. Beware the state licensure regulation may require medical air pipe system with a certain number of outlets for the functional space of the patient care room. FGI Construction Guidelines for Health Care Facilities has been adopted by most health care facility regulators at the state level.